Attenuation unit, dBm or dB/cm 回答済み
Hello,
I have some trouble in defining the material properties and the unit for damping (attenuation). Should it be dB/m or dB/cm? I know that PZFlex is a user-defined unit system and I can work in any parameter system that I prefer. I defined all my parameters in SI unit and just was not sure about attenuation. I have three materials as air, water and LDPE. Water and LDPE attenuation at 1 MHz are 0.217 dB/m and 400 dB/m. My operating frequency is frefund = 6 MHz. I defined the materials in this way
matr
wvsp on
type tisu
prop watnl 1000 1482 0 0.01 0.0 0.0 5 -2e+08 2e+08
sdmp watnl $frefund db 0.217 0 1e+06 2
thrm watnl 4180 0.6 0.6 0.6
wvsp on
type elas
prop LDPE 930 2080 0 0.01
vdmp LDPE $frefund db 400 0 1e+06 1
thrm LDPE 2100 0.33 0.33 0.33
wvsp on
type elas
prop air 1.24 343 0
end
Did I define the attenuation correctly or I should change the attenuation for water and LDPE to dB/cm as 0.00217 dB/cm and 4 dB/cm, respectively? Also, did I define the power-law power loss by sdmp and vdmp correctly (my operating frequency is frefund = 6 MHz and the attenuation values of 0.217 dB/m and 400 dB/m are at 1 MHz)?
Thank you for your support.
Best,
Ahmad
5件のコメント
Hello Ahmad,
Since OnScale's default units are meters you should define attenuation in dB/m as you have in your example.You have also defined the power-law power loss correctly. I have changed a few things in your code, mainly in your watnl definition - bovera pmax pcut must be defined on a new line. This is not currently described in the command reference which we will change.
matr
wvsp on
type tisu
prop watnl 1000 1482 0 0.01 0.0 0.0
5 -2e+08 2e+08
sdmp watnl 6e6 db 0.217 0 1e6 2
thrm watnl 4180 0.6 0.6 0.6
end
matr
wvsp on
type elas
prop LDPE 930 2080 0 0.01
vdmp LDPE 6e6 db 400 0 1e+06 1
thrm LDPE 2100 0.33 0.33 0.33
end
matr
wvsp on
type elas
prop air 1.24 343 0
end
Best Regards,
Team OnScale.
Thanks Chloe. It was a great help. In case I run the script in PZFlex on my PC (not OnScale), is my definition of attenuation in dB/m still correct?
Thanks again.
Ahmad
Ahmad,
PZFlex also works in dB/m so you can use this material definition.
Best Regards,
Team OnScale.
Thanks for your response.
I tried to run the model using
symb #get { step } timestep
symb simtime = 2 * ( ( $x3 - $x2 ) / $velmin )
symb nexec = nint ( $simtime / $step )
symb nloops = 40
symb nexec_loop = nint ( $nexec / $nloops )
proc plot save
exec $nexec_loop
I adjusted the function to be a continuous sine wave and the simulation time as 2 * ( ( $x3 - $x2 ) / $velmin ) to just consider for the wave travel inside material 2, which is (x3-x2), reach the back surface and get reflected until the reflected wave reaches the front surface of this material. I derived the loss and used that in the thermal script to derive the temperature rise. As I increased the simulation time (e.g. 3 * ( ( $x3 - $x2 ) / $velmin ), or 4 * ( ( $x3 - $x2 ) / $velmin ) ), I expected to see more temperature rise in material 2 as more energy was deposited. As I expected, the loss was increased by increasing the simtime, but the temperature rise in material 2 decreased by increasing the simtime. Would you please let me know if it is reasonable to get lower temperature rise, while the loss is increasing? Is it possible to send you the acoustic and thermal codes?
Thank you for your help.
Ahmad
Typically, the loss will accumulate under CW drive so we would expect temperature to build increase over time. Thermal simulations are usually a 2 stage model: acoustic model to determine the losses (simulation typically in milliseconds) for X number of cycles and a separate temperature simulation to apply the loses over a long simulation period (typically in seconds)
If you are only using a single model, what you may be seeing is that your temperature simulation is not running for any substantial amount of time to capture any meaningful increase in temperature.